What radiation detector can tell energy level?

In summary, radiation detectors such as GM counters, scintillation detectors, and proportional counters are useful for detecting and determining the energy of specific types of radiation. For spectroscopy and determining energy levels of radionuclides, detectors that measure energy, such as scintillation detectors or Ge(Li) detectors, are required. The energy level of the radiation is proportional to the number of electrons released from the scintillator, which is influenced by the intensity of the light emitted. The resulting pulse height spectra may have multiple photopeaks with accompanying Compton tails, making analysis challenging. The efficiency of scintillation detectors can range from 2 to 50% depending on various factors.
  • #1
DetectiveT
9
1
Hello everyone. Is there a type of radiation detector that can tell the energy level? I have a basic understanding about Geiger counter and Scintillation counter, they seem only tell the incident particle number.
The testing environment may contain Thorium series, Uranium series and Actinium series decay chain. The purpose is trying to find out what particles are there, or at least what decay (energy level) is in there mainly. Please let me know what kind of radiation detectors suits most. Thank you!
 
Physics news on Phys.org
  • #2
Radiation detectors are sensitive to and useful for specific types or radiation. For example a GM counter is useful for detecting x-rays , gamma rays or beta particles above a certain energy but cannot tell you the energy of the radiation. A scintillation detector is useful for detecting and determining the energy of x-rays , gamma rays and beta particle above certain energies. A proportional counter can also determine energies of specific types or radiation.

For determining the energy levels of specific radionuclides, that is doing spectroscopy, requires the identification and determination of the energy of the radiation. Detective work in determining how this radiation is related to those energy level. This require detectors that measure energy e.g. scintillation detectors or Ge(Li) detectors.

In the decay of radionuclides the radiation you detect is from the transitions from some higher state to a lower state (not necessarily the ground state) . So basically you are measuring the energy difference of the excited states. If the decay is to the ground state then you have the energy level of that excited state. But often the decay is though an intermediate state(s) resulting in two or more gamma rays. You have to figure out how to combine these energies of the radiations to get energy levels consistent with the gamma rays you measure.
 
  • Like
Likes mfb
  • #3
gleem said:
Radiation detectors are sensitive to and useful for specific types or radiation. For example a GM counter is useful for detecting x-rays , gamma rays or beta particles above a certain energy but cannot tell you the energy of the radiation. A scintillation detector is useful for detecting and determining the energy of x-rays , gamma rays and beta particle above certain energies. A proportional counter can also determine energies of specific types or radiation.

For determining the energy levels of specific radionuclides, that is doing spectroscopy, requires the identification and determination of the energy of the radiation. Detective work in determining how this radiation is related to those energy level. This require detectors that measure energy e.g. scintillation detectors or Ge(Li) detectors.

In the decay of radionuclides the radiation you detect is from the transitions from some higher state to a lower state (not necessarily the ground state) . So basically you are measuring the energy difference of the excited states. If the decay is to the ground state then you have the energy level of that excited state. But often the decay is though an intermediate state(s) resulting in two or more gamma rays. You have to figure out how to combine these energies of the radiations to get energy levels consistent with the gamma rays you measure.

Thank you very much! I understand. May I ask a further question regarding to the scintillation detector please. If my understanding is correct, the measured energy level is proportional to the electrons released from the scintillator, and the scintillator works based on photoelectric effect, which each of the incident photon interacts with a single electron, then how does a photon with higher energy level ejects more electrons than a low energy incident photon in order to give a higher reading? Thank you.
 
  • #4
DetectiveT said:
Thank you very much! I understand. May I ask a further question regarding to the scintillation detector please. If my understanding is correct, the measured energy level is proportional to the electrons released from the scintillator, and the scintillator works based on photoelectric effect, which each of the incident photon interacts with a single electron, then how does a photon with higher energy level ejects more electrons than a low energy incident photon in order to give a higher reading? Thank you.

The radiation interacts with the scintillation material (SM) via all possible EM interactions with the Compton effect dominating exciting the atoms in the SM with a subsequent emission of some light. The SM must be transparent to this light to allow it to be collected by the photo multiplier tube (PMT). The light photons entering the PMT cause the emission of electrons from a photo cathode the number of which depends on the intensity of that light which in turn depends on the energy deposited in the SM. The PMT amplifies the charge liberated by its photo cathode to produce a usable signal for further electronic processing. The signal strength depends on the energy deposited in the SM. Below is a typical pulse height spectra of a single gamma ray

upload_2016-2-4_9-51-14.png


Characteristically it has a peak representative of the total energy deposited in the SM called the photopeak along with a fairly uniform tail extending to the lowest pulse height representing whose processes that only leave a part of the radiation energy in the SM. This tail is primarily due to the Compton effect and the subsequent degrading of the energy of those electrons released.

In a spectrum of a more complex decay you will see many photopeaks with their accompanying Compton tails superimposed. The full width at half max of the peak is typically about 8% of the pulse height. It can be challenging to analyze these spectra.

Efficiencies of scintillation detectors are typically 2 to 50 % depending on energy, detector size and source detector distance.
 

Attachments

  • upload_2016-2-4_9-49-57.png
    upload_2016-2-4_9-49-57.png
    21.1 KB · Views: 465
  • upload_2016-2-4_9-50-34.png
    upload_2016-2-4_9-50-34.png
    12.9 KB · Views: 437
  • #5
DetectiveT said:
and the scintillator works based on photoelectric effect, which each of the incident photon interacts with a single electron, then how does a photon with higher energy level ejects more electrons than a low energy incident photon in order to give a higher reading? Thank you.
You get a whole chain of processes in the material. The photon gives most or all of its energy to an electron, this electron excites various other electrons, it can also lead to the emission of more photons of intermediate energy, which again transfer their energy to more electrons, those excite more electrons and can emit more photons, and so on.

The excited electrons in the scintillating material then emit low-energetic photons - typically in the visible range, sometimes infrared or UV - where the material is transparent. Those photons get detected.
You always have some fluctuations in the number of scintillation photons, this limits the energy resolution of the detector.
 
  • #6
gleem said:
The radiation interacts with the scintillation material (SM) via all possible EM interactions with the Compton effect dominating exciting the atoms in the SM with a subsequent emission of some light. The SM must be transparent to this light to allow it to be collected by the photo multiplier tube (PMT). The light photons entering the PMT cause the emission of electrons from a photo cathode the number of which depends on the intensity of that light which in turn depends on the energy deposited in the SM. The PMT amplifies the charge liberated by its photo cathode to produce a usable signal for further electronic processing. The signal strength depends on the energy deposited in the SM. Below is a typical pulse height spectra of a single gamma ray

View attachment 95282

Characteristically it has a peak representative of the total energy deposited in the SM called the photopeak along with a fairly uniform tail extending to the lowest pulse height representing whose processes that only leave a part of the radiation energy in the SM. This tail is primarily due to the Compton effect and the subsequent degrading of the energy of those electrons released.

In a spectrum of a more complex decay you will see many photopeaks with their accompanying Compton tails superimposed. The full width at half max of the peak is typically about 8% of the pulse height. It can be challenging to analyze these spectra.

Efficiencies of scintillation detectors are typically 2 to 50 % depending on energy, detector size and source detector distance.

Thank you very much! I do understand now, it is really helpful and detailed, thank you!:smile:
 
  • Like
Likes gleem
  • #7
mfb said:
You get a whole chain of processes in the material. The photon gives most or all of its energy to an electron, this electron excites various other electrons, it can also lead to the emission of more photons of intermediate energy, which again transfer their energy to more electrons, those excite more electrons and can emit more photons, and so on.

The excited electrons in the scintillating material then emit low-energetic photons - typically in the visible range, sometimes infrared or UV - where the material is transparent. Those photons get detected.
You always have some fluctuations in the number of scintillation photons, this limits the energy resolution of the detector.

I see, thank you as well!
 

Related to What radiation detector can tell energy level?

1. How does a radiation detector measure energy level?

A radiation detector measures energy level by detecting the amount and type of radiation that is emitted from a source. Different types of radiation, such as alpha, beta, and gamma rays, have different energy levels and can be detected by specific types of detectors.

2. What types of radiation detectors can measure energy level?

There are several types of radiation detectors that can measure energy level, including Geiger counters, scintillation detectors, and ionization chambers. Each type has its own method of detecting and measuring radiation energy.

3. Can a radiation detector measure the energy level of all types of radiation?

No, a radiation detector may only be able to measure the energy level of specific types of radiation, depending on its design and capabilities. Some detectors may be able to measure multiple types of radiation, while others are specialized for a specific type.

4. How accurate are radiation detectors in measuring energy level?

The accuracy of a radiation detector in measuring energy level depends on various factors, such as the type of detector, the sensitivity of the instrument, and the calibration of the equipment. In general, modern radiation detectors are highly accurate and can measure energy levels with precision.

5. Can a radiation detector measure the energy level of non-ionizing radiation?

No, most radiation detectors are designed to measure ionizing radiation, which has enough energy to strip electrons from atoms. Non-ionizing radiation, such as radio waves and visible light, does not have enough energy to be detected by these instruments. Specialized detectors may be used to measure non-ionizing radiation, such as photodiodes for visible light or microwave detectors for radio waves.

Similar threads

  • High Energy, Nuclear, Particle Physics
Replies
6
Views
1K
  • High Energy, Nuclear, Particle Physics
Replies
6
Views
803
  • High Energy, Nuclear, Particle Physics
Replies
1
Views
110
Replies
1
Views
413
  • Introductory Physics Homework Help
Replies
3
Views
110
  • High Energy, Nuclear, Particle Physics
Replies
1
Views
1K
  • High Energy, Nuclear, Particle Physics
Replies
3
Views
1K
  • High Energy, Nuclear, Particle Physics
Replies
9
Views
4K
  • Nuclear Engineering
Replies
25
Views
4K
  • High Energy, Nuclear, Particle Physics
Replies
14
Views
2K
Back
Top