Design a system to measure the electroretinogram (ERG)

In summary: What filter topology and component values have you chosen for your anti-alias filter and why?So in summary, you are working on designing a differential amplifier circuit to measure the electroretinogram. You have chosen resistor values and filter topology for the anti-alias filter. You have also chosen a sample rate for the ADC.
  • #1
benderlee
5
1
Homework Statement
Design a system to measure the electroretinogram (ERG). Use a differential amplifier to maximize the CMRR, and condition the signal appropriately for a 16-bit A/D system with input range ±10 Volts. Give all component values.

Suggest an appropriate sampling rate, assuming that all noise has been sufficiently attenuated by the time the signal reaches the digitizer, and give the resolution of the signal in V/bit at the digitizer.


ERG bandwidth= 0-300 Hz

ERG dynamic range = ± 1 mV
Relevant Equations
(See attached image)
243239
 
Physics news on Phys.org
  • #2
Welcome to the PF. :smile:

Thanks for posting some of the relevant equations. Now please show us your work on the solution. Also, what do the sensors for an ERG look like, and where are they placed?
 
Last edited:
  • Like
Likes benderlee
  • #4
benderlee said:
I am truly stumped by this one, looking for more information online myself and was seeing if anyone in this great community knows anything about this, so far I found this:

https://webvision.med.utah.edu/book/electrophysiology/the-electroretinogram-erg/
That's a helpful link. So it looks like one ERG electrode is placed on the cornea, and the other is attached to the patient's body for reference.

So start with the desired sampling rate (how is it based on the signal bandwidth?), and design the differential amplifier circuit and other gain elements and filter elements next. Are you familiar with designing digitizing systems, including anti-alias filters? That is all part of this problem you are working on.

You can also probably do some reading about ECG and EEG measurement systems, since they will share a lot of the same design components (they will have different bandwidth and pickup voltages, but the rest is pretty similar).
 
  • Like
Likes benderlee
  • #5
So this is how the actual circuit would be set up 😊
 

Attachments

  • erg.JPG
    erg.JPG
    33.2 KB · Views: 348
  • #6
benderlee said:
So this is how the actual circuit would be set up 😊
Did you draw that, or find it somewhere? A differential amp has two signal inputs, not three.
 
  • Like
Likes benderlee
  • #7
This is a diagram from a textbook - attached is another diagram from another text. that is more or less the same with a ground, reference, and corneal electrode:
 

Attachments

  • erg2.JPG
    erg2.JPG
    45.8 KB · Views: 371
  • #8
Also FYI: in the ERG, the differential amp records the difference between the active electrode (attached to the cornea) and the reference electrode (attached to the outer canthus). The ground electrode is attached to the forehead.
 
  • #9
benderlee said:
This is a diagram from a textbook
That's not a very accurate or informative diagram, unfortunately. Hopefully you are learning about differential amplifier design for biomedical applications from a better and more detailed textbook or application notes.
benderlee said:
Also FYI: in the ERG, the differential amp records the difference between the active electrode (attached to the cornea) and the reference electrode (attached to the outer canthus). The ground electrode is attached to the forehead.
So the diff amp inputs should come from closely spaced electrodes at the eye, and the "ground"/reference connection comes from some other large part of the body nearby. The "ground"/reference connection goes into the overall amplifier circuit, not into the diff amp itself.

So show us what you have figured out so far. What resistor values have you chosen for your diffamp and why? What filter topology and component values have you chosen for your anti-alias filter and why?

What sample rate have you chosen for your ADC and why?
 
  • Like
Likes benderlee

Related to Design a system to measure the electroretinogram (ERG)

1. What is an electroretinogram (ERG)?

An electroretinogram (ERG) is a diagnostic test that measures the electrical activity of the retina in response to light stimulation. It is used to assess the function of the retina and diagnose various eye diseases.

2. How does an ERG work?

An ERG works by placing electrodes on the surface of the eye to measure the electrical signals produced by the retina in response to light stimulation. The signals are then amplified and recorded, allowing for the assessment of the retina's function.

3. What are the components of a system to measure ERG?

A system to measure ERG typically includes a light source to stimulate the retina, electrodes to record the electrical signals, an amplifier to amplify the signals, and a recording device to capture the signals for analysis.

4. How is an ERG test performed?

An ERG test is performed by first dilating the patient's pupils and then placing electrodes on the surface of the eye. The patient is then exposed to flashes of light, and the electrical signals produced by the retina are recorded and analyzed.

5. What are the benefits of using an ERG system?

An ERG system can provide valuable information about the function of the retina and help diagnose various eye diseases. It is a non-invasive and relatively quick test, making it a useful tool for monitoring the progression of eye diseases and assessing the effectiveness of treatments.

Similar threads

  • Engineering and Comp Sci Homework Help
Replies
8
Views
483
  • Engineering and Comp Sci Homework Help
Replies
13
Views
415
  • Engineering and Comp Sci Homework Help
Replies
4
Views
861
  • Engineering and Comp Sci Homework Help
Replies
5
Views
893
  • Engineering and Comp Sci Homework Help
Replies
5
Views
1K
  • Introductory Physics Homework Help
Replies
3
Views
185
  • Engineering and Comp Sci Homework Help
Replies
1
Views
677
  • Engineering and Comp Sci Homework Help
Replies
12
Views
1K
  • Science and Math Textbooks
Replies
5
Views
1K
  • Engineering and Comp Sci Homework Help
Replies
4
Views
977